# Using PCR plates on STAR

**URL:** <https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74>\
**Category:** resources\
**Created:** [November 21, 2024, 2:59am UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74 "2024-11-21T02:59:57Z")\
**Posts on this page:** 12\
**Page:** 2

<div class="post-metadata">

**Author:** ![CamilloMoschner](https://discuss.pylabrobot.org/user_avatar/discuss.pylabrobot.org/camillomoschner/32/7_2.png) [@CamilloMoschner](https://discuss.pylabrobot.org/u/CamilloMoschner)\
**Post date:** [December 9, 2024, 11:05pm UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/21 "2024-12-09T23:05:21Z")

</div>

> [@rickwierenga](#):
>
> can you use the adapter if it is one of the ones in PLR. @CamilloMoschner do you recognize the adapter Danylo is using?

No, I have never seen this orange plate\_adapter before.

@dkone Is this the deck setup you use for the images you’ve shown?

> [@mackenzie](#):
>
> ```auto
> pcr_plt_car_0 = PLT_CAR_L5PCR(name='pcr_plate_carrier_0')
> pcr_plt_car_0[0] = pcr_adapter_00 = Hamilton_96_adapter_188182('pcr_adapter_00')
> pcr_adapter_00.plate_z_offset=3.2
> 
> ```
> 
> Next, trying to set the 3.2 mm value when assigning the plate to the adapter (which is not the correct way to go about it, as it’s setting the entire plate above the adapter, but wanted to see if this hack worked)
> 
> ```auto
> pcr_plate_00 = Eppendorf_96_wellplate_250ul_Vb_semiskirted('pcr_plate_00')
> pcr_adapter_00.assign_child_resource(pcr_plate_00, location=Coordinate(0, 0, 3.2))
> 
> ```

If yes, then I can help you fix the **defining plate adapter** issue because you are using the wrong plate\_adapter definition for your specific, new-to-PLR plate\_adapter.

As background:

 ![Hamilton_96_adapter_188182](https://discuss.pylabrobot.org/uploads/default/original/1X/890a0af83e04b4c3b2a45567c635e69c584b1caa.jpeg)  
This is the Hamilton plate adapter from the definition you have posted above. It is `size_x=110.0` and `size_y=75.0`, i.e. not SLAS/ANSI 1-2004 (R2012), i.e. x=127.76 mm x y=85.48 mm - which your orange, plastic plate\_adapter appears to be.  
If you have used this definition, and you set it not into the x-y center of the plate\_site but to the plate\_site _origin_, i.e. front-left-bottom,  
(which is what `pcr_plt_car_0[0] = pcr_adapter_00 = Hamilton_96_adapter_188182('pcr_adapter_00')` does)  
then this explains why your channels move too far to the left.

Another example of a plate\_adapter (to showcase a small fraction of the diversity of plate\_adapters) that I hope to add to the PLR Resource Library soon is

 ![Screenshot 2024-12-09 at 23.15.30](https://discuss.pylabrobot.org/uploads/default/original/1X/86a542cc4f79d8b66e27e3b86fb345d0b05680ed.png)  
The 96-wellplate aluminium adapter does have SLAS/ANSI 1-2004 footprint in this case.  
And this seems a lot closer to the orange, plastic plate\_adapter in your images.

Do you have a manufacturer catalogue number of your plate\_adapter?  
With that we might be able to find some engineering drawings for it to base a new definition on.

---

<div class="post-metadata">

**Author:** ![dkone](https://discuss.pylabrobot.org/letter_avatar_proxy/v4/letter/d/b4bc9f/32.png) [@dkone](https://discuss.pylabrobot.org/u/dkone)\
**Post date:** [December 9, 2024, 11:27pm UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/22 "2024-12-09T23:27:23Z")

</div>

I am using a PCR tube rack which fits really well onto the plate carrier.  
For the code I don’t define an adapter. I use `plt_car[1] = Thermo_FS_96_wellplate_300ul_Vb_semi_skirted(name='source') ` It was working fine a month ago. Currently I am creating a new Thermo Fisher Scientific plate.  
I can provide dimensions of the adapter.

---

<div class="post-metadata">

**Author:** ![dkone](https://discuss.pylabrobot.org/letter_avatar_proxy/v4/letter/d/b4bc9f/32.png) [@dkone](https://discuss.pylabrobot.org/u/dkone)\
**Post date:** [December 9, 2024, 11:31pm UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/23 "2024-12-09T23:31:32Z")

</div>

[https://www.azerscientific.com/Adapter-For-0.2-ML-PCR-Strips-Microplate-Centrifuge](https://www.azerscientific.com/Adapter-For-0.2-ML-PCR-Strips-Microplate-Centrifuge)  
This adapter looks very similar to the one I am using. I think it is this one.

---

<div class="post-metadata">

**Author:** ![CamilloMoschner](https://discuss.pylabrobot.org/user_avatar/discuss.pylabrobot.org/camillomoschner/32/7_2.png) [@CamilloMoschner](https://discuss.pylabrobot.org/u/CamilloMoschner)\
**Post date:** [December 10, 2024, 12:46am UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/24 "2024-12-10T00:46:08Z")

</div>

> [@dkone](#):
>
> Eppendorf\_96\_wellplate\_250ul\_Vb and decided to use it. For some reason the tips landed around that black wire

> [@dkone](#):
>
> For the code I don’t define an adapter. I use `plt_car[1] = Thermo_FS_96_wellplate_300ul_Vb_semi_skirted(name='source')`

Right, this makes sense too:

Problem 1:  
`Eppendorf_96_wellplate_250ul_Vb` → [link to docs](https://docs.pylabrobot.org/resources/library/eppendorf.html)  
This plate has a footprint of 123x81 mm^2, i.e. also not ANSI/SLAS 1-2004 of 127.76x85.48 mm^2.  
But the plate\_holder/\_site on a Hamilton carrier is ANSI/SLAS 1-2004 of 127.76x85.48 mm^2.

You assigned 123x81 mm^2 to the front-left-bottom of the carrier plate\_holder → everything is shifted to the left in your defintion as a result, explaining why your tips landed around the black wire.

Problem 2:  
Using a semi-skirted plate without a plate\_adapter.  
@rickwierenga, I think we should raise an error when people assign semi-skirted and non-skirted plates to anything else but a `PlateAdapter()`.  
Physically speaking these plates cannot simple sit on a flat surface by themselves… a gust of wind will change their x-y position.

This also causes issues with your well\_cavity\_bottom → physically speaking your wells are now raised in the z-dimension, but your machine doesn’t know that because it is not being told that the wellplate sits on top of a plate\_adapter with a `plate_adapter.plate_z_offset` associated with the plate-plate\_adapter combo.

Solution: model reality

1. Generate new plate definition
2. Generate new plate\_adapter definition
3. Measure specific plate-plate\_adapter `plate_z_offset`

I will make a lot of infographics next week to give some visual guidance 🙂

---

<div class="post-metadata">

**Author:** ![dkone](https://discuss.pylabrobot.org/letter_avatar_proxy/v4/letter/d/b4bc9f/32.png) [@dkone](https://discuss.pylabrobot.org/u/dkone)\
**Post date:** [December 10, 2024, 12:58am UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/25 "2024-12-10T00:58:39Z")

</div>

Thank you for such a detailed response @CamilloMoschner!  
I am working on creating and testing new plate. I might need help with `_compute_height_from_volume_Thermo_FS_96_wellplate_300ul_Vb_semi_skirted`  
The orange adapter is a standard PCR tube rack. Could we add it to the resources? I am also thinking that people could use standard PCR strip tubes on this adapter?

 ![IMG_6159](https://discuss.pylabrobot.org/uploads/default/original/1X/d1dc347a1861f67adf86b524f28ddb50d92cdc97.jpeg)  
 ![IMG_6158](https://discuss.pylabrobot.org/uploads/default/original/1X/063a720dadfa759eaef8073f7f1493538061df96.jpeg)

---

<div class="post-metadata">

**Author:** ![CamilloMoschner](https://discuss.pylabrobot.org/user_avatar/discuss.pylabrobot.org/camillomoschner/32/7_2.png) [@CamilloMoschner](https://discuss.pylabrobot.org/u/CamilloMoschner)\
**Post date:** [December 10, 2024, 1:00am UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/26 "2024-12-10T01:00:09Z")

</div>

Note to avoid confusion (poka-yoke):  
ANSI/SLAS 1-2004 of 127.76x85.48 mm^2 is the exact footprint dimensions as defined by this plate standard.  
Most machines, including our favs from Hamilton, do not allow firmware commands with a precision exceeding 0.1mm.  
As a result, you might see some plate definitions as 127.8x85.5 mm^2, rounded to the closest 0.1 mm.

I tend to still give the precise 127.76x85.48 mm^2 for completeness sake, and because PLR will automatically round to the closest 0.1 mm anyway before sending commands to your machine.  
Plus, we might use more precise machines in the future of PLR 🙂

---

<div class="post-metadata">

**Author:** ![rickwierenga](https://discuss.pylabrobot.org/user_avatar/discuss.pylabrobot.org/rickwierenga/32/963_2.png) [@rickwierenga](https://discuss.pylabrobot.org/u/rickwierenga)\
**Post date:** [December 10, 2024, 1:35am UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/27 "2024-12-10T01:35:21Z")

</div>

> [@CamilloMoschner](#):
>
> @rickwierenga, I think we should raise an error when people assign semi-skirted and non-skirted plates to anything else but a `PlateAdapter()`.

_anything_ else? i disagree with that. But an error when assigning to a PlateHolder would probably be good.

wdyt [PlateHolder only takes non-skirted plates by rickwierenga · Pull Request #330 · PyLabRobot/pylabrobot · GitHub](https://github.com/PyLabRobot/pylabrobot/pull/330)

---

<div class="post-metadata">

**Author:** ![CamilloMoschner](https://discuss.pylabrobot.org/user_avatar/discuss.pylabrobot.org/camillomoschner/32/7_2.png) [@CamilloMoschner](https://discuss.pylabrobot.org/u/CamilloMoschner)\
**Post date:** [December 10, 2024, 7:57am UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/28 "2024-12-10T07:57:09Z")

</div>

I’m a bit confused:  
Shouldn’t the PR be “PlateHolder only takes _skirted_ Plates or PlateAdapters or TubeRacks”?

---

<div class="post-metadata">

**Author:** ![rickwierenga](https://discuss.pylabrobot.org/user_avatar/discuss.pylabrobot.org/rickwierenga/32/963_2.png) [@rickwierenga](https://discuss.pylabrobot.org/u/rickwierenga)\
**Post date:** [December 10, 2024, 9:14pm UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/29 "2024-12-10T21:14:50Z")

</div>

yes smh

---

<div class="post-metadata">

**Author:** ![rickwierenga](https://discuss.pylabrobot.org/user_avatar/discuss.pylabrobot.org/rickwierenga/32/963_2.png) [@rickwierenga](https://discuss.pylabrobot.org/u/rickwierenga)\
**Post date:** [December 10, 2024, 9:17pm UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/30 "2024-12-10T21:17:09Z")

</div>

if it is a plate, we check that it is skirted. for other items we don’t check since the user may have unpredictable deck setups and I think we shouldn’t introduce artificial constraints.

---

<div class="post-metadata">

**Author:** ![CamilloMoschner](https://discuss.pylabrobot.org/user_avatar/discuss.pylabrobot.org/camillomoschner/32/7_2.png) [@CamilloMoschner](https://discuss.pylabrobot.org/u/CamilloMoschner)\
**Post date:** [December 11, 2024, 12:50am UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/31 "2024-12-11T00:50:11Z")

</div>

> [@dkone](#):
>
> I am working on creating and testing new plate. I might need help with `_compute_height_from_volume_Thermo_FS_96_wellplate_300ul_Vb_semi_skirted`

@dkone, what would you like to use the height computation for, is this for liquid level following during aspiration/dispensation?

> [@dkone](#):
>
> The orange adapter is a standard PCR tube rack. Could we add it to the resources?

Yes, absolutely!  
I recommend having a look at the Hamilton plate\_adapter definition, copying it, and modifying it for your new plate\_adapters geometry.  
Then we just need to know the manufacturer and part number to add it to the PLR Resource Library.

> [@dkone](#):
>
> I am also thinking that people could use standard PCR strip tubes on this adapter?

haha yeah 🙂  
I have done the exact same thing with Opentrons OT-2 API a long time ago:  
In PLR there are multiple ways we can do this but I still think the way I did it back then is the most robust.  
Once this plate\_adapter has been added to the PLR Resource Library you can just use essentially any plate definition as a proxy for your strip-tubes, and just adjust the `plate_adapter.plate_z_offset` to what you need it to be.  
This will allow you to use all positions very quickly.

(The alternative would be to model the orange thing not as a `PlateAdapter` but as a `TubeRack`. But once you do that you cannot use a plate on it anymore [I believe]. Therefore I recommend generating a `PlateAdapter` definition for it, and using a plate definition for the strip-tubes)

---

<div class="post-metadata">

**Author:** ![dkone](https://discuss.pylabrobot.org/letter_avatar_proxy/v4/letter/d/b4bc9f/32.png) [@dkone](https://discuss.pylabrobot.org/u/dkone)\
**Post date:** [December 12, 2024, 10:47pm UTC](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74/32 "2024-12-12T22:47:05Z")

</div>

> [@CamilloMoschner](#):
>
> @dkone, what would you like to use the height computation for, is this for liquid level following during aspiration/dispensation?

Yes, I want to use liquid level following for this plate.

> [@CamilloMoschner](#):
>
> Therefore I recommend generating a `PlateAdapter` definition for it, and using a plate definition for the strip-tubes

Will create a `PlateAdapter` then.

[Previous page](https://discuss.pylabrobot.org/t/using-pcr-plates-on-star/74.md?page=1)