Hamilton STAR Dispensing Inaccuracies

Hi All,

I am a new user to both Hamilton Systems (Have used Opentrons a lot in the past) and pyLabRobot. For context I am also more of a molecular biologist then automation engineer, but have been tasked with protocol development for our STAR.

I wrote a normalization protocol that theoretically should variable volume repeat dispense diluent into an empty PCR plate then add the correct amount of sample into each plate. I was running water tests and was seeing variable volume levels at the end of the run. After further testing, I figured out that the first part (repeat dispensing diluent) was adding too much water to each well. When I fixed the concentrations to all be the same, it should have been adding 36uL but there was ~39-40 uL in each well.

I was not using a default hamilton liquid class, but I listed the settings I was using below:
flow rate: 100
aspiration settling time: 0.3
dispense settling time: 1
aspiration height: Using LLD
dispense height: fixed height of 1mm

Does anyone have any insight on if this is a liquid class issue or a hardware issue? The system was purchased used, but validated by a third party distributer. I am happy to share more details or the protocol itself if that would be helpful. Thank you all in advance!

One way to know if the instrument is in spec is to run the FVK (field verification kit). Hamilton uses scales to verify with a special liquid that your instrument is in spec.

I have no idea what those parameters you are using but typically I start with the default water then tune it gravimetrically with a scale. Elevation and humidity play a large role and liquid classes need to be tuned. Once tuned we fined in the same environment they transfer pretty dang well instrument to instrument.

1 Like

Thank you that makes sense. After more digging I think the volume correction curve as causing me my issues. When I changed it from the default from the liquid class to a flat curve all my measurements were coming in more accurately. Some fine tuning is definitely still needed, but it is performing much better. I don’t currently have any scales that can connect with the hamilton, but I will look into it.

Thanks again for the insight!